Hello Xavior,
-First go to the device manager, if com3 is not there, ad manually a com3 port.( it is not really PnP)
-Then go to the properties of this port and find the screen where you can select the IRQ and Adress.
-set to configuration 5
-Now click on the irq (probably not 5 but some thing else at this moment) and change it to 5
Restart windows and check if it is still there.
Do not start monitor/download
Go to c:\lasercad
Run LBSinst.exe
There you should see LBS setting ( I do not remember exactly but LPT1 and com3 should be there)
Now be careful All your needed configuration data about Field correction and Galvo timings are still in the flash memory of the LBS card!!
After pressing install or something on the LBSinst.exe you get the question where to get the configuration.
-From PC
-From laser
-New
Choose from Laser.
Now it should work.
You can set Monitor in the startup of windows. Then download start's also
When all is working you turn key and press shutter open in monitor.
The picture with the plug I talked about is the back of the cabinet
You see there also the monitor/keyboard plug and a smal D9 for a start stop interface
Best regards,
Martin
PS this setting should not be made,