give me a call, i can give you more info over the phone.
Alex.
[email protected]
Type: Posts; User: CT Tech
give me a call, i can give you more info over the phone.
Alex.
[email protected]
ECDuplex...You ever get this fixed?
if you need help. send me an email. Be more than happy to help you over the phone.
take care.
Alex
The generator rarely gets reagjusted. You do this when you incoming power is really bad. I mean bad, machine shuts down when incoming power is more than 3 % of what the transformers are set to.
And...
Check your incoming power. Make sure that yor internal transformers are set to the same as your 3 phase. You cannot have no more that 3 % difrence between leggs R-S-T and no more than 4% diffrence...
Well this is very simple. if you shut the machine down, and your bsatteries are dead. Well, i believe its time to replace them assuming their not leaking.
The batteries when brand new, can save your...
You have switches on the door for these options. You can select ascii, text, Binary, etc.
You also have your baud rate otions by these switches.
your default setting are.
baud rate is 4800
data...
Teh 290 ,300,310,500,510 with Charmilles control ,have this intialize option.
These machines are initilized, by holding down one of the buttons on the control panel. The greeen cycle start key, or...
Let me know if you need any more help, or your problem is solved.Alex.
Don't know if your problem was fixed. but their are ways to figure it out. email me if you still have questions. [email protected]
Alex
I don't know if this was solved. But just to let you know. Their is 2 contacts on that switch in the fron door, Upper left if the door is the sliding type, and upper right if it's the swing type. ...
If you still need help.Email me [email protected]. Alex